digital clipping at 8240 input - Forum
Genelec Community Forum has been archived
Thank you for all the years of Community discussions and activity!
The time has come for us to retire the community forums, but we will keep everything available for reading. There is a lot of valuable content written over the years, and you'll be able to access all of that. However, no new posts can be written, or old posts modified.
If you have questions, we recommend you use the Support function on the bottom right corner to contact our Customer Support directly.
Alternatively, if you wish to engage in a community discussion with other people, there are many active forums available. There is also an active, fan managed GENELEC COMMUNITY in Facebook, and many Genelec employees are active in the discussions there.
We are sorry for any possible inconvenience this may cause you, but we hope to hear from you through the other channels mentioned above.
Genelec Support
Message Boards
digital clipping at 8240 input
dumbo, modified 13 Years ago.
digital clipping at 8240 input
Padawan Posts: 78 Join Date: 12/20/09 Recent PostsI'd very much appreciate any help on this matter.
Ps The Canadian Genelec "expert" repeatedly told me that, to solve this problem, I needed to reduce the volume at the GLM interface (even after I told him this had no effect whatsoever).
christophe-anet, modified 13 Years ago.
Re: digital clipping at 8240 input
Jedi Knight Posts: 188 Join Date: 3/23/09 Recent PostsThanks for your post. Here is an answer from Dr Aki Mäkivirta:
This is normal. The 8240A has a detection of the AES digital audio data level right at the AES input. If the input level exceeds -0.1 dB FS the red clip LED turns on immediately for a small period of time.
The purpose of this is to safeguard the engineer while mixing or recording to make sure he/she does not inadvertently overload the digital bus. Commercial CDs have typically been maximized and this may push the peaks very close or exactly at 0 dB FS, turning on the clip LED. While this is distracting it does not indicate a problem.
The solution to turn off the red LED is to attenuate the ingoing digital audio bit stream by at least 0.1 dB. This function is usually available in an audio workstation or similar source of the digital audio signal.
A hardware AES/EBU volume control doing this is also offered by Studio Technologies Inc. (http://www.studio-tech.com/), and it connects between the loudspeakers and the digital audio source.
Best regards,
Genelec Oy
Aki Mäkivirta
R&D Manager for Digital Products
dumbo, modified 13 Years ago.
Re: digital clipping at 8240 input
Padawan Posts: 78 Join Date: 12/20/09 Recent Postsdumbo, modified 13 Years ago.
I don't think I explained my original problem properly.
Padawan Posts: 78 Join Date: 12/20/09 Recent PostsHello,
Thanks for your post. Here is an answer from Dr Aki Mäkivirta:
This is normal. The 8240A has a detection of the AES digital audio data level right at the AES input. If the input level exceeds -0.1 dB FS the red clip LED turns on immediately for a small period of time.
The purpose of this is to safeguard the engineer while mixing or recording to make sure he/she does not inadvertently overload the digital bus. Commercial CDs have typically been maximized and this may push the peaks very close or exactly at 0 dB FS, turning on the clip LED. While this is distracting it does not indicate a problem.
The solution to turn off the red LED is to attenuate the ingoing digital audio bit stream by at least 0.1 dB. This function is usually available in an audio workstation or similar source of the digital audio signal.
A hardware AES/EBU volume control doing this is also offered by Studio Technologies Inc. (http://www.studio-tech.com/), and it connects between the loudspeakers and the digital audio source.
Best regards,
Genelec Oy
Aki Mäkivirta
R&D Manager for Digital Products
I've been thinking about this and wondering if the issue isn't a bit more complicated. When the room correction filters are being generated in Autocal it appears that, at some frequencies, some gain is applied. With other room correction systems (TactAudio for example) this gain runs the risk of causing digital clipping of of fully modulated CDs (at that the frequencies where the gain has been applied). The solution to this is to attenuate the filter curves. I suspect this might be a problem with the 8240s because I'm getting the overload indicators flashing all the time and I it's hard to believe that so many recordings have be mastered with the peaks clipped (digitally). Is this possibly whats going on and, if so, is there any way to attenuate the filters?
1038, modified 13 Years ago.
Re: digital clipping at 8240 input
Jedi Master Posts: 340 Join Date: 4/6/09 Recent PostsThere is NO gain applied during EQ, only attenuation.
There are numerous CD's that always illuminate the predicive clip light when i'm playing at low spl levels. This occurence is not model specific, i see it on 8240, 8250's and 8260's.
The amount of music that has been produced way too hot is scary

Rest assured though that no gain is applied by autocal.
Cheers
Steve
maza, modified 11 Years ago.
Re: digital clipping at 8240 input
Youngling Posts: 1 Join Date: 3/11/12 Recent Postsi have also this kind of problem. My 8240's are clipping red when in Ableton Live master is around -4.5db.
I have tried to change the glm volume but that doesn't seem to have any affect. None of my lines in ableton are clipping and the master either, but still the indicator on the speakers turns red when i put the master around -4.5db. Any idea what is causing this? Could it be my sound card or something?
Thanks in advance!
ilkka-rissanen, modified 11 Years ago.
Re: digital clipping at 8240 input
Yoda Posts: 2564 Join Date: 3/23/09 Recent PostsThe clipping is caused by a too high audio signal from your sound card. The red light indicates that the level of the incoming signal is too high and that´s why the LED still flashes red although you turn down the level from GLM. Let´s check a few things to start with; are you feeding your speakers with a digital or analog signal? If it is digital, clipping should not occur as it is at a fixed level. If it is analog and your sound card has an external “master volume” pot for the outgoing level you should turn it down until the LED stops indicating clipping of the input signal and then instead increase volume from GLM. Hope these tips help!